Otii Ace Pro

Hardware overview

The front side of the Otii Ace Pro has the main connectors, additional ports to extend measurement capabilities, plus a status LED.

Ace front

The back side of the Otii Ace Pro has the host USB connector and an input socket to power up the unit with an external DC adapter when required.

Ace back

Main connections

The Otii Ace Pro has the voltage(+) and voltage(-) binding post (red and black connectors located on the right side of the instrument), which are used to source the DUT but also to measure the current, voltage, and power of the device. The DUT can be connected through its battery connectors or DC input; refer to the connection diagram below for better understanding.

Ace connected to DUT

To power up the Otii Ace Pro and enable the communication with the host computer, use the provided USB C cable to connect the unit to your computer.

Ace connected to a computer

Check out the Wiring up page to explore all possible connection configurations between the Otii Ace Pro and your DUT.

Expansion Port

The expansion port on the front of the Otii Ace Pro is designed to enhance the hardware's functionality, allowing users to measure additional voltages and currents or trigger external events. These additional pins feature multi-input and multi-output for analog and digital signals, serial communication, and additional power and ground points. Below is a general overview of the additional pins available:

Connector Description Maximum rating
0-15V Adjustable output between 0V and 15V 600 mA
DGND Digital ground (return current)
RX UART RX / Digital Input (Exclusively under Scripting) 0 V to 5.5 V
TX UART TX / Digital Output (Exclusively under Scripting) 0 V to 5.5 V
GPI1 Digital input 1 0 V to 5.5 V
GPI2 Digital input 2 0 V to 5.5 V
GPO1 Digital Output 1 0 V to 5.5 V
GPO2 Digital Output 2 0 V to 5.5 V
DGND Digital ground (return current)
AGND Analog ground (return current) / Analog measurements
ADC+ Differential input for current measurement. It also measures single ended voltage with respect to AGND -10 V to 25 V
ADC- Input for current measurement -10 V to 25 V
SENSE+ Analog input with the AGND as reference/return -10 V to 25 V
SENSE- Analog input with the AGND as reference/return -10 V to 25 V

Status LED

The table below explains the Otii Ace Pro's status LED behaviors and their meanings to help you quickly identify the device's state.

LED behavior Otii Ace Pro Status
Solid blue Otii Ace Pro is active
Solid green Relay closed, sinking/sourcing power or in-line mode
Flashing white Calibrating
Flashing red/blue Upgrading firmware

Ace Tech spec

General

Operating environment Min Typical Max
Temperature 10°C / 50°F 30°C / 86°F
Humidity 30 % 60 %

Main

Power supply Min Typical Max
Output voltage 0 V 25 V
Output voltage setting resolution 1 mV
Self-consumption 3.5 W
Output power, max continuous 30 W 10 14
Output power, max peak 50 W 10
Voltage between USB/DC jack and Main – -200 V -200 V
Voltage between DGND/AGND and Main – -200 V -200 V
Programmable current sink (requires an Otii Toolbox) Min Typical Max
Sink current 0 A 5 A
Sink current, setting resolution 1 µA
Sink voltage 0 V 25 V
Sink power, max continuous 15 W
Sink power, max peak 125 W
Current measurement Min Typical Max
Accuracy -5A to 5 A ±(0.05 % + 25 nA)
Resolution 0.4 nA
Sample rate 50 ksps
Analog bandwidth (3 dB) 50 kHz
Voltage measurement Min Typical Max
Accuracy ±(0.01 % + 1 mV)
Sample rate 50 ksps
Analog bandwidth (3 dB) 50 kHz
Output voltage readback resolution 3.5 µV

Expansion port

UART Min Typical Max
Bitrate 50 bps 5.25 Mbps
Digital I/O Min Typical Max
Digital IO operating voltage 1.2 V Vio 15 5 V
VIL Low-level input voltage Vio * 0.2 V
VIH High-level input voltage Vio * 0.8 V
Imax, Max sink/source current (total for GPIOs) -10 mA 10 mA
Differential ADC, pins ADC- and ADC+ Min Typical Max
Voltage input -10 V 25 V
Shunt voltage range -102.4 mV 102.4 mV
Resolution 12.2 nV
Accuracy ±(0.1 % + 1 µV)
Input impedance > 100 Mohm
Single ended ADC, pins ADC- and ADC+ Min Typical Max
Voltage input -10 V 25 V
Resolution 3.1 µV
Accuracy ±(0.1 % + 250 µV)
Input impedance > 100 Mohm
SENSE, pins SENSE- and SENSE+ Min Typical Max
Voltage input -10 V 25 V
Resolution 3.1 µV
Accuracy ±(0.1 % + 250 µV)
Input impedance > 100 Mohm
Expansion port power supply Min Typical Max
Output voltage 0 V 15 V
Output voltage setting resolution 5 mV
Output current 600 mA
Voltage between USB/DC jack & DGND/AGND -200 V 11 12 -200 V 11 12

USB and DC jack

DC jack Min Typical Max
Input voltage 7 V 20 V
Input current 5 A
USB Min Typical Max
VBUS voltage 13 4.75 V 20 V
VBUS current 13 3 A