Otii Arc Pro

Hardware overview

The front side of the Otii Arc Pro has the main connectors, additional ports to extend measurement capabilities, plus a status LED.

Arc front

The back side of the Otii Arc Pro has the host USB connector and an input socket to power up the unit with an external DC adapter when required.

Arc back

Main connections

The Otii Arc Pro has the voltage(+) and voltage(-) binding post (red and black connectors located on the right side of the instrument), which are used to source the DUT but also to measure the current, voltage, and power of the device. The DUT can be connected through its battery connectors or DC input; refer to the connection diagram below for better understanding.

Arc connected to DUT

To power up the Otii Arc Pro and enable communication with the host computer, connect the unit to your computer using the provided USB micro B to USB A cable.

Arc connected to a computer

Check out the Wiring up page to explore all possible connection configurations between the Otii Arc Pro and your DUT.

Expansion Port

The expansion port on the front of the Otii Arc Pro is designed to enhance the hardware's functionality, allowing users to measure additional voltages and currents or trigger external events. These additional pins feature multi-input and multi-output for analog and digital signals, serial communication, and additional power and ground points. Below is a general overview of the additional pins available:

Connector Description Maximum rating
+5V 5 V output 500 mA
DGND Digital ground (return current)
RX UART RX / Digital Input (Exclusively under Scripting) 0 V to 5.5 V
TX UART TX / Digital Output (Exclusively under Scripting) 0 V to 5.5 V
GPI1 Digital input 1 0 V to 5.5 V
GPI2 Digital input 2 0 V to 5.5 V
GPO1 Digital Output 1 0 V to 5.5 V
GPO2 Digital Output 2 0 V to 5.5 V
DGND Digital ground (return current)
AGND Analog ground (return current) / Analog measurements
ADC+ Differential input for current measurement. It also measures single ended voltage with respect to AGND 0 V to 5.5 V
ADC- Input for current measurement 0 V to 5.5 V
SENSE+ Analog input with the AGND as reference/return 0 V to 5.5 V
SENSE- Analog input with the AGND as reference/return 0 V to 5.5 V

Status LED

The table below explains the Otii Arc Pro's status LED behaviors and their meanings to help you quickly identify the device's state.

LED behavior Otii Arc Pro Status
Flashing green Otii Arc Pro is active
Solid green Relay closed, sinking/sourcing power

Arc Tech spec

General

Operating environment Min Typical Max
Temperature 10°C / 50°F 30°C / 86°F
Humidity 30 % 60 %
USB power supply 1 Min Typical Max
Output voltage (auto range) 0.5 V 3.75 V
Output voltage (locked to high current range) 0.5 V 4.2 V
Output voltage setting resolution 1 mV
Output current, max continuous 3 up to 1 A 1 2.4 A
Output current, max peak 3 up to 1 A 1 5 A
External 7.5 v - 9 v power supply 2 Min Typical Max
Output voltage (auto range) 0.5 V 4.55 V
Output voltage (locked to high current range) 0.5 V 5.0 V
Output voltage setting resolution 1 mV
Output current, max continuous 3 2.4 A
Output current, max peak 3 5 A
Programmable current sink (requires an Otii Toolbox) Min Typical Max
Sink current 0 A 2.5 A
Sink current, setting resolution 39 µA
Sink voltage, USB power supply 0.85 V 4 4.2 V
Sink voltage, external power supply 0.85 V 4 5.0 V
Current measurement Min Typical Max
Accuracy ±(0.1 % + 50 nA)5
Sample Rate in ±19 mA range 4 ksps
Sample Rate in ±2.7A range 1 ksps
Sample Rate in ±5.0 A range 1 ksps
Analog bandwidth (3 dB) 400 Hz
Voltage measurement Min Typical Max
Total accuracy ±(0.1 % + 1.5 mV)
Sample rate 1 ksps

Expansion port

UART Min Typical Max
Bitrate 110 bps 5.25 Mbps
Digital I/O 6 Min Typical Max
Digital IO operating voltage 1.2 V Vio 7 5 V 8
VIL Low-level input voltage Vio * 0.2 V
VIH High-level input voltage Vio * 0.8 V
Imax, Max sink/source current (total for GPIOs) -10 mA 10 mA
Differential ADC, pins ADC- and ADC+ 9 Min Typical Max
Voltage input 0 V 5 V
Shunt voltage range -81.9175 mV 81.2 mV
Resolution 2.5 µV
Accuracy ±(0.1 % + 10 µV)
Input impedance 220 kΩ
Single ended ADC, pins ADC- and ADC+ 9 Min Typical Max
Voltage input 0 V 5 V
Resolution 1.25 mV
Accuracy ±(0.1 % + 7.5 mV)
Input impedance 830 kΩ
SENSE, pins SENSE- and SENSE+ Min Typical Max
Voltage input 0 V 5 V
Resolution 1.5 mV
Accuracy 1 %
Input impedance 1 MΩ
Expansion port power supply Min Typical Max
Output voltage 5 V 10
Output current 500 mA

USB and DC jack

DC jack Min Typical Max
Input voltage 7 V 9 V
Input current 5 A
USB Min Typical Max
VBUS voltage 4.75 V 5.25 V
VBUS current 3 A